Here's WPI
Remember samples are blinded and from someone who tested positively and this is newly drawn, never before stored blood at the WPI
They wrongly identified a negative control as positive - but think they can explain that..
They were able to identify XMRV correctly in 3 of 4 positives - it wasn't easy, they could only do so on one day and they missed on one but the important thing is that by and large they identified the positive samples correctly - a big win
This suggests a) because the controls tested negative, their lab is not contaminating the samples
b) while its not always easy to find XMRV they can find it and in mostly the right samples but not always....
On the other hand if they had only 5 samples - they would have known 4 should have been positive - so much for blinding...and on one of the days they misidentified which one was positive.
It's a little sketchy because they missed on one control and XMRV is not easy to find - and, suggests that really, no one has a really good test for it yet.
The salient fact is though, that, by and large the WPI picked out the controls from the positives and that is the acid test. It suggests to me that the WPI is doing some right the other labs are not and that the BWG should be very carefully going over the labs procedures - down to the smallest detail - to see what the WPI is doing differently.
To me this seems this is something like a weak win for the WPI. If they had gotten that negative right it would have been bigger.
If you put this in a chart you find that they picked the right positive only 2/5 times..They got the negative right 1/2 times....They didn't even mention in this chart what happened to samples 2,3,4 on Day 0!
The straight PCR test needs alot of work - which is why the WPI is now focusing on culturing I suppose.
Phase IIb WPI results
Nested RT-PCR for MRV gag followed by sequencing of positive bands to confirm specificity
-RNA extracted directly from plasma
-Total nucleic acid extracted from PBMC and RT step performed
WB testing has not been completed (error discovered)
PBMC results:
-For Day 0, Subject 1 and the pedigreed negative control were positive (note on slide: investigation following decoding of results determined that there was a procedural error during PBMC sample extraction involving reuse of needles, employed to lyse cells and shear DNA, on sequential PBMC cell pellets.)
-For Day 2 only, Subjects 1 and 4 were negative as well as the pedigreed negative control
-For Day 2 only, Subjects 2 and 3 were positive
Plasma results: All subjects and the control were negative for both days